02-07-2025, 12:11 PM
I have owned this instrument from new (c1990) and it has always worked flawlessly. Until now. It's stuck in some kind of boot loop. It says "memory test" on screen for a few seconds, then boots to a normal display and is responsive to controls for about a second or two. Then goes back to the memory test. Rinse and repeat. I don't recall ever seeing the memory test screen before.
I have the original user and maintenance manuals as supplied. They don't contain schematics, not even a decent block diagram. There is a list of modules, their functions and a few photos showing their locations. Unlike many Tektronix manuals, there is no faultfinding guide.
Looking at this video: https://www.youtube.com/watch?v=0VhyDC1DTCk&t=36s it looks like there's a lithium thionyl chloride battery in there which is probably now dead. The manual mentions EEPROMs so it's not clear if calibration data is stored in battery backed RAM or not.
Any ideas? I'm off to the Broadcast Engineering Museum for a few days so won't be looking a the 4195A immediately.
I have the original user and maintenance manuals as supplied. They don't contain schematics, not even a decent block diagram. There is a list of modules, their functions and a few photos showing their locations. Unlike many Tektronix manuals, there is no faultfinding guide.
Looking at this video: https://www.youtube.com/watch?v=0VhyDC1DTCk&t=36s it looks like there's a lithium thionyl chloride battery in there which is probably now dead. The manual mentions EEPROMs so it's not clear if calibration data is stored in battery backed RAM or not.
Any ideas? I'm off to the Broadcast Engineering Museum for a few days so won't be looking a the 4195A immediately.