19-05-2015, 01:48 PM
It looks like my earlier hunch was promising 
With 0.1V across R138, that gives us a tail current of 0.66mA. In other words, nothing at all. So why is this?
Only 0.9V across R138. Hmm... I don't know what the tail current should be, and the manual doesn't seem to give any clues, so let's pretend for a moment that it should be 10mA (it might be double that in practice). In which case, there should be 1.5V across R138, and so 2.1 or 2.2V across R138. As it stands, TR82 is lacking in bias. So that points to R137 or R152.
I'd put money on one (or both) of them being high in value or maybe completely O/C. You could try measuring the resistance in circuit, with the 'scope powered down for a couple of minutes first. If you get a lower than expected reading, then that's in-circuit conditions causing currents to flow thought a different path, but if you get a higher than expected reading, then the resistor is faulty.
If you don't want to remove the PCB or CRT, you might be able to change them from the component side if there is enough space. Just snipe away the old component, leaving enough of the old leads poking through the PCB, and solder the replacements to that. Do it quickly, otherwise the heat will conduct through and melt the original solder on the rear of the PCB. It's a bit of a bodge, but it does work. You can always "make good" the repair should another fault force the complete strip-down at some point in the future...
Regarding the PSU rails, they're not regulated in this model, so the numbers given will always be "nominal". But the low reading might be indicative of another fault - we'll perhaps learn more later.

With 0.1V across R138, that gives us a tail current of 0.66mA. In other words, nothing at all. So why is this?
Only 0.9V across R138. Hmm... I don't know what the tail current should be, and the manual doesn't seem to give any clues, so let's pretend for a moment that it should be 10mA (it might be double that in practice). In which case, there should be 1.5V across R138, and so 2.1 or 2.2V across R138. As it stands, TR82 is lacking in bias. So that points to R137 or R152.
I'd put money on one (or both) of them being high in value or maybe completely O/C. You could try measuring the resistance in circuit, with the 'scope powered down for a couple of minutes first. If you get a lower than expected reading, then that's in-circuit conditions causing currents to flow thought a different path, but if you get a higher than expected reading, then the resistor is faulty.
If you don't want to remove the PCB or CRT, you might be able to change them from the component side if there is enough space. Just snipe away the old component, leaving enough of the old leads poking through the PCB, and solder the replacements to that. Do it quickly, otherwise the heat will conduct through and melt the original solder on the rear of the PCB. It's a bit of a bodge, but it does work. You can always "make good" the repair should another fault force the complete strip-down at some point in the future...
Regarding the PSU rails, they're not regulated in this model, so the numbers given will always be "nominal". But the low reading might be indicative of another fault - we'll perhaps learn more later.