22-05-2015, 11:33 AM
Firstly, I wouldn't worry too much about the transistor substitutions at this stage. Manufacturers frequently made changes - mostly to save costs, or perhaps improve reliabilty in light of experience gained. Often, seemingly unusual choices might have been made because it meant using fewer different types overall. And every designer has their favourites.
Now we have more faults to explore. The various symptoms are interesting, to say the least, and it's difficult to know which to tackle next.
OK, so still have dots, so no sweep. Are you sure the controls are set correctly for a sweep? The fact that channel 2's Y-shift control moves the dots diagonally suggests that you might be in X-Y mode, but of course there might be a fault. So, can we set the controls as follows:
Ensure the 'scope is in non-storage mode, which I think should be just a case of making sure the "STORE" button is out.
Press "REP" so that the timebase is not in single-shot mode.
Ensure all 7 of the "TRIG MODE" buttons are out. This should put the 'scope into auto-trigger mode, meaning it free-runs in the absence of a valid trigger signal.
Although this 'scope has a "STABILITY" control, and if this is turned fully anti-clockwise, the sweep will stop. Turn it fully clockwise.
Set the timebase "TIME/DIV" to somewhere in the middle. 1ms/div should be fine.
Make sure it's not in X-Y mode. It looks like you have to pull out the red "VARIABLE" timebase knob on this one.
Press in the red knob on the X position control so it's not in X5 magnification mode...
As for the Y channels, just set those to something like 1V/DIV, position in the middle, and to be honest, and set to ALT if you want both traces on.
Having done all that, have you got a sweep?
How experienced are you with 'scopes generally? The trouble is, even with a fully working unit, it can be quite hard to get a stable and sensible display. When fault-finding a 'scope, you really do need to draw on operational experience. If you tell me your level, then hopefully I can occupy the safe ground between "teaching you to such eggs" and "losing you completely"
Cheers,
Mark
Now we have more faults to explore. The various symptoms are interesting, to say the least, and it's difficult to know which to tackle next.
OK, so still have dots, so no sweep. Are you sure the controls are set correctly for a sweep? The fact that channel 2's Y-shift control moves the dots diagonally suggests that you might be in X-Y mode, but of course there might be a fault. So, can we set the controls as follows:
Ensure the 'scope is in non-storage mode, which I think should be just a case of making sure the "STORE" button is out.
Press "REP" so that the timebase is not in single-shot mode.
Ensure all 7 of the "TRIG MODE" buttons are out. This should put the 'scope into auto-trigger mode, meaning it free-runs in the absence of a valid trigger signal.
Although this 'scope has a "STABILITY" control, and if this is turned fully anti-clockwise, the sweep will stop. Turn it fully clockwise.
Set the timebase "TIME/DIV" to somewhere in the middle. 1ms/div should be fine.
Make sure it's not in X-Y mode. It looks like you have to pull out the red "VARIABLE" timebase knob on this one.
Press in the red knob on the X position control so it's not in X5 magnification mode...
As for the Y channels, just set those to something like 1V/DIV, position in the middle, and to be honest, and set to ALT if you want both traces on.
Having done all that, have you got a sweep?
How experienced are you with 'scopes generally? The trouble is, even with a fully working unit, it can be quite hard to get a stable and sensible display. When fault-finding a 'scope, you really do need to draw on operational experience. If you tell me your level, then hopefully I can occupy the safe ground between "teaching you to such eggs" and "losing you completely"

Cheers,
Mark